我有一个带模拟调制选项的Agilent N5181A。
作为良好实践,我的测试以“仪器预设”开始。 这将显示器置于其默认频率/幅度显示模式。 当启用幅度调制时,会在显示屏上显示,但没有%深度,信号源,音频波形/频率等的指示。如果我暂停测试,请点击“本地”按钮,然后点击'AM '键,具有所有所需细节的AM菜单显示在频率/幅度显示下方,并在我继续测试时持续存在(带有频率和幅度的进一步命令)。 有没有办法用N5181A的母语显示AM菜单? 看来如果我命令N5181A模拟其他一些信号发生器之一,我可能会这样做,但我还没来得及玩这个。 然后我需要切换回N5181A的母语继续测试。 AM菜单会持续存在吗? 我的替代方案是保持仿真模式并在我的测试程序中更改为模拟信号发生器的相应驱动程序。 以上来自于谷歌翻译 以下为原文 I have an Agilent N5181A with the analog modulation option. As good practice my tests begin with 'Instrument Preset'. This puts the display into its default frequency/amplitude display mode. When amplitude modulation is enabled, this is indicated on the display, but there is no indication of % depth, source, audio waveform/frequency, etc. If I pause my test, hit the 'Local' button, and then hit the 'AM' key, the AM menu with all the desired details appears below the frequency/amplitude display and persists when I continue the test (with it further commands of frequency and amplitude). Is there a way to display the AM menu in the native language of the N5181A? It appears I might be able to do this if I command the N5181A to emulate one of number of other signal generators, but I have not had the time (yet) to play around with this. I would then need to switch back to the native language of the N5181A to continue the test. Would the AM menu persist? My alternative is to stay in emulation mode and change to the appropriate driver in my test program for the emulated signal generator. |
|
相关推荐
1个回答
|
|
Hi -Have MXG模拟另一个信号发生器不会改变仪器PRESET设置.N5181A可以按照您希望的方式设置AM;
那么你必须将它保存到仪器状态。 然后,作为程序中开启程序的一部分,您必须首先调用仪器状态和AM; 然后N5181A将按照您的设置配置显示器。这些按键保存仪器状态(根据需要设置仪器后):保存 - >选择Reg: - > 01(或您想要的任何数字) - >保存Reg :这些击键将调用仪器状态和AM软键:Recall - > Select Reg: - > 01 - > Enter - > AMRegards - 以上来自于谷歌翻译 以下为原文 Hi - Having the MXG emulate another signal generator will not change the instrument PRESET settings. The N5181A can be setup the way you want it to be with AM; then you have to save it to an instrument state. Then as part of your turn-on routine in your program you will have to first recall the instrument state and AM; then the N5181A will have the display configured as you had set it up. These keystrokes save the instrument state (after setting the instrument as desired): Save -> Select Reg: -> 01 (or whatever number you want) -> Save Reg: These keystrokes will recall the instrument state and AM softkeys: Recall -> Select Reg: -> 01 -> Enter -> AM Regards - |
|
|
|
只有小组成员才能发言,加入小组>>
1342 浏览 0 评论
2400 浏览 1 评论
2227 浏览 1 评论
2088 浏览 5 评论
2980 浏览 3 评论
1282浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
788浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
1074浏览 0评论
1342浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2025-1-26 05:09 , Processed in 1.182236 second(s), Total 75, Slave 59 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号