几个月前我买了一台二手MSO7104A。
它最后一次校准是在2009年,我注意到较低的伏/格设置有一些偏移问题,所以我决定运行一个用户cal。 我按照手册中的指示,使用三个BNC三通将BNC电缆从触发输出连接器连接到前面板。 但是,当我运行用户cal时,它在运行不到30秒后失败,其中“Failure:Digitizer,failed,init for talon cal,setup chan1”。 范围似乎工作正常,它通过自检。 我尝试了几种不同的布线方式,结果相同。 有谁知道这个错误意味着什么,有没有办法完成校准或回滚校准数据? 以上来自于谷歌翻译 以下为原文 I bought a used MSO7104A a few months ago. It was last calibrated in 2009 and I have noticed some offset issues on the lower volts/div settings, so I decided to run a user cal. I connected the BNC cables from the trigger out connector to the front panel with three BNC tees as directed in the manual. However, when I run the user cal, it fails after running for less than 30 seconds with "Failure: Digitizer, failed, init for talon cal, setup chan1". The scope seems to work fine and it passes the self test. I have tried several different cabling arrangements with the same result. Does anyone know what this error means and if there is any way to either complete the calibration or roll back the calibration data? |
3个回答
错误消息表示设备上的A / D未正确与内存控制器通信。
即使该装置通过自检,用户校准在这方面也更为严格。 这是不可恢复的。 您需要将其退回维修。 短发 以上来自于谷歌翻译 以下为原文 The error message means that the A/D on your unit is not communicating with the memory controller properly. Even though the unit passes self-test, the user calibration is more rigorous in this area. This is not recoverable. You will need to return it for repair. Bob |
如果错误位于A / D的低位,则在操作期间可能看不到任何可见问题。
使用默认校准系数时,设备将起作用,但可能不符合规格。 短发 以上来自于谷歌翻译 以下为原文 If the error is in the lower bits of the A/D, you may not see any visible issues during operation. With default calibration factors, the unit will function, but it may not meet specifications. Bob |
你看到的尖峰是你失败的原因。
为什么在正常操作中没有看到它们仍然取决于A / D中位故障的位置。 cal以与正常操作略有不同的模式运行A / D,以发现这些类型的问题并确保硬件正常运行。 以上来自于谷歌翻译 以下为原文 The spikes you see are the cause of your failure. Why you don't see them in normal operation is really still dependent on where the bit failure is in the A/D. The cal runs the A/D in slightly different modes than normal operation to uncover these types of issues and insure that the hardware is functioning properly. |
1033 浏览 0 评论
2278 浏览 1 评论
2065 浏览 1 评论
1937 浏览 5 评论
2808 浏览 3 评论
801浏览 1评论
588浏览 0评论
684浏览 0评论
2466浏览 0评论
1523浏览 0评论
GMT+8, 2024-9-21 10:53 , Processed in 1.159520 second(s), Total 47, Slave 42 queries .
