嗨,我有一台HP8596E频谱分析仪,有以下故障排除方法: - 输入电平:CAL OUT(300 MHz,-20 dBm) - 在所有_even_位置(-20,-40和-60 dB)上切换输入衰减器时,
测得的电平是正确的(-20 dBm)。 只是本底噪声水平正在增加,这是正常的。 - 在所有_odd_位置(-10,-30,-50和-70 dB)上切换输入衰减器时,测量电平错误(约-15至-20 dB),但仅在10至15秒内。 在这段时间之后,水平缓慢增加并再次变得正确。 有人对这种行为有所了解以及如何治愈这种行为? 我会很感激一些提示。 关心多米尼克 以上来自于谷歌翻译 以下为原文 Hi, I have a HP8596E spectrum analyzer with the following troubleshooting: - Input level: CAL OUT (300 MHz, -20 dBm) - When switching the input attenuator on all its _even_ positions (-20, -40 and -60 dB), the measured level is correct (-20 dBm). Just the noise floor level is increasing, which is normal. - When switching the input attenuator on all its _odd_ positions (-10, -30, -50 and -70 dB), the measured level is wrong (about -15 to -20 dB), but only during 10 to 15 seconds. After this delay of time, the level increases slowly and becomes correct again. Does somebody have an idea about this behavior and how to cure that? I would appreciate some hints. Regards Dominique |
Hi Dominique - 输入衰减器内部有(3)个内部焊盘部分,总共有70 dB的输入衰减;
10 dB部分,20 dB部分和40 dB部分。 例如,为了实现30 dB的输入衰减,接入20 dB和10 dB部分。由于20 dB,40 dB和60 dB设置似乎正常运行,因此每次显示10 dB部分有缺陷 接通时,存在振幅问题。 输入衰减器部件号(33321-60037)不再可用,因为产品超出支持寿命几年。 安捷伦仍在销售较低级别的部件,如10 dB电路,触点等。 如果您愿意打开输入衰减器,您可能能够找出哪个部件有缺陷。 这可能具有挑战性,因为输入衰减器本身没有示意图。 99%的时间交换机逻辑是正确的,问题的原因是硬件本身。 当10 dB部分出现问题时,通常表示功率过大的情况。 也许可以看到视觉损坏,可以订购和更换适当的下层部件。 问候 - 以上来自于谷歌翻译 以下为原文 Hi Dominique - There are (3) internal pad sections inside the input attenuator that make up a total of 70 dB of input attenuation; a 10 dB section, 20 dB section and 40 dB section. To achieve 30 dB of input attenuation for example, the 20 dB and 10 dB sections are switched in. Since the 20 dB, 40 dB and 60 dB settings appear to be operating correctly, it appears the 10 dB section is defective because every time that is switched in, there is an amplitude problem. The input attenuator part number (33321-60037) is no longer available as the product is several years beyond the support life. Lower level parts such as the 10 dB circuit, contacts, etc. are still sold by Agilent. If you’re willing to open up the input attenuator, you may be able to find out which part is defective. This may be challenging as there are no schematic diagrams for the input attenuator itself. 99% of the time the switch logic is correct and the cause of the problem is the hardware itself. When the 10 dB section has a problem it is usually indicative of an over power situation. Perhaps visual damage can be seen and appropriate lower level parts can be ordered and replaced. Regards - |
醉狼工作室 发表于 2019-1-15 11:25 您好Tabbott,非常感谢您的解释。 我将尝试打开10 dB部分,看看有什么损坏。 修好后(我希望如此),我会随时通知你。 此致,多米尼克 以上来自于谷歌翻译 以下为原文 Hi Tabbott, Thank you very much for your explanation. I will try to open the 10 dB section and see what is damaged. When repaired (I hope it), I will keep you informed. Regards, Dominique |
1394 浏览 0 评论
2424 浏览 1 评论
2275 浏览 1 评论
2112 浏览 5 评论
3027 浏览 3 评论
1488浏览 1评论
837浏览 0评论
1252浏览 0评论
1394浏览 0评论
GMT+8, 2025-3-6 03:51 , Processed in 1.341022 second(s), Total 46, Slave 41 queries .