在几天的反复试验过程中,我发现当频率选项设置为CONT时,从前面板记录一系列频率测量失败(使用AUTO和REC工作正常)。
无论门的设置如何,记录都将继续进行,保存999个正确值,然后快速填充所需数量的测量值的剩余部分,值为9.91E + 37。 该计数器最初具有固件版本1.10,但行为未更改,更新为1.12。 通过Agilent Interactive IO使用远程SCPI命令不会发生同样的问题。 使用该方法,日志记录正常工作。 有什么建议,或者我应该简单地将此报告为错误? 在远离办公室/实验室工作时,只使用计数器和USB密钥记录大量无间隙频率数据的能力是仪器的一个重要特征。 没有这项工作是令人失望的。 标记 以上来自于谷歌翻译 以下为原文 Over the course of a couple of days of trial and error I found that logging a series of frequency measurements from the front panel failed when the frequency option was set to CONT (worked fine with AUTO and REC). Regardless of the setting of the gate, the logging would proceed, save 999 correct values, then rapidly fill the remainder of requested number of measurements with values of 9.91E+37. The counter originally had firmware revision 1.10, but the behavior was unchanged with updating to 1.12. The same issue did not occur using remote SCPI commands via Agilent Interactive IO. Using that method, the logging worked properly. Any suggestions, or should I simply report this as a bug? The ability to log large amounts of gap-free frequency data using only the counter and a USB key is a valuable feature of the instrument when working away from the office/lab. Not having this work is a disappointment. Mark |
|
相关推荐
3个回答
|
|
连续模式存在缺陷,使得测量超时设置应用于样品计数(持续时间)中的所有读数,而不是每个读数。
因此,如果超时在达到读取持续时间之前到期,则所有后续读数将具有9.91E + 37(非数字)指定。 要在连续模式下防止这种情况,请按:实用程序 - > Instr设置 - >超时设置 - >超时(关)或者,设置一个超过计数中所有读数的累计门时间的超时值。 此缺陷已得到修复,将在下一次固件更新中提供。 此版本的发布日期尚未确定。 以上来自于谷歌翻译 以下为原文 There is a defect in Continuous mode such that the measurement timeout setting is applied to ALL readings in the sample count (duration), rather than to each individual reading. Thus, if the timeout expires before the reading duration is reached, all subsequent readings will have the 9.91E+37 (not a number) designation. To prevent this in Continuous mode, press: Utilty -> Instr Setup -> Time Out Settings -> Time Out (Off) Or, set a timeout value much larger than the cumulative gate time for all readings in the count. This defect has been fixed and will be available in the next firmware update. The date for this release has not been determined. |
|
|
|
子虚道人 发表于 2018-12-25 08:24 谢谢。 建议的解决方法是有效的。 不确定我是否可以在这个问题上得到“二合一”,但我注意到CONT设置的频率测量有时无法锁定。这里的解决方法是循环到AUTO(立即采集)和 再次回到CONT。 再次,在使用SCPI远程命令时,我没有注意到这种行为。 补救措施很简单,但我们将不胜感激。 以上来自于谷歌翻译 以下为原文 Thank you. The suggested workaround is effective. Not sure if I can get a "two-for-one" on this question, but I've noticed that frequency measurements with the CONT setting sometimes fail to lock in. The workaround here is to cycle to AUTO (acquisition is immediate) and back to CONT again. Again here I haven't noticed this behavior when using remote commands with SCPI. The remedy is simple, but a fix would be appreciated. |
|
|
|
hzp_bbs 发表于 2018-12-25 08:37 实际上,你可以得到两个一个。 此行为也是已解决但未在固件更新中发布的缺陷。 此修复将出现在与CONT模式修复中的测量超时相同的版本(发布日期TBD)中。 以上来自于谷歌翻译 以下为原文 Actually, you can get two for one. This behavior is also a defect that has been resolved, but not released in a firmware update. This fix will appear in the same release (release date TBD) as the measurement timeout in CONT mode fix. |
|
|
|
只有小组成员才能发言,加入小组>>
1356 浏览 0 评论
2409 浏览 1 评论
2239 浏览 1 评论
2093 浏览 5 评论
2984 浏览 3 评论
1324浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
802浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
1111浏览 0评论
1356浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2025-2-3 15:52 , Processed in 1.199579 second(s), Total 45, Slave 40 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号